I am wanting to use these inside fridge and freezer compartments to log door opening habits of occupants for a fridge replacement program.
Associated Product 

Comments & Subscriptions


 

Thanks for the post.
This could work for this application however keep in mind a couple of limitations:
The LCD may not operate correctly at temperatures below 0C. This will not affect the logging of the data however.
This logger is not condensation-proof so, as the door opens and the warm air hits the cold logger, condensation may form on the logger's electronic components which could lead to premature logger failure.
Also, at temperatures below the specified logger range, battery life may be shorter than specified.
Happy logging!