MX2501 pH data logger resistant to greasy acid media?

I'd like to know, whether the MX2501 pH data logger is resistant to greasy acid media. The application is a the discharge of a grease filter. pH range 4..6 pH (4000-8000 microSiemens), temperature 0-50°C. Does anyone have experience with a similar application?
Associated Product 

Comments & Subscriptions


 

Thanks for the post!. We have not tested the MX2501 product in this type of environment therefore we would not recommend using it that way. The MX2501 is intended to monitor pH in a bodys of water. Lakes, rivers, wells, applications like that. The electrode will get fouled enough by the oil that it will likely inhibit the ability for the electrode to accurately report the pH of the sample to the logger. I hope this helps.